T-2420 SPECIFICATIONS

3350 Scott Boulevard, Building No. 32
Santa Clara, California 95054-3118
Phone: 408.496.6838
Fax: 408.496.6918

E-Mail: [email protected]
URL: http://www.thermonics.com

Specifications subject to change without notice.

©

6/96 Thermonics

Printed in U.S.A.

GENERAL DESCRIPTION

Method of Cooling: Forced air with single-stage mechanical refrig-
eration and multistage self-cooling heat exchanger.

Method of Heating: Forced air with in-line air heaters.

Operator Interface: Twenty character alphanumeric display and
keypad; menu-driven software with self-diagnostics.

Controls: Microprocessor control of all critical functions.

PHYSICAL DESCRIPTION

Model #

Height

Width

Depth

Weight

(in/cm)

(in/cm)

(in/cm)

(lbs/kgm)

T-2420

40/102

24/61

21/53

290/132

T-2420 with
TF-1C5 or TF-IC7

48/123

24/61

49/124

355/161

POWER REQUIREMENTS

Model #

Volts (AC)

Hertz

Phases

Amps

T-2420

208

23050

/60 1

φ

20

TEMPERATURE PERFORMANCE

Model #

Range

Accuracy

Stability

Displayed

Transition

(°C)*

(°C)

(°C)

Resolution(°C)

Time (sec)

T-2420 with
T-BOH

-75

–200

±1.0

±0.3

0.1

150

T-2420SX with
T-BOH

-85

–225

±1.0

±0.3

0.1

120

T-2420 with
TF-IC5

-75

–205

±1.0

±0.3

0.1

90

T-2420 with
TF-IC7

-75

–225

±1.0

±0.3

0.1

5

T-2420SX with
TF-IC7

-85

–225

±1.0

±0.3

0.1

4

FEATURES

Calibration: Automatic or manual. Up to 10 different applications may be cali-
brated independently and stored in non-volatile memory.

Temperature Sensor: Device under test (DUT), K-type thermocouple. Optional
T-type thermocouple. 1000 ohm RTD for auto calibration.

Controller: Microprocessor-based PID (Proportional, Integral, Differential) with
user control of DUT mass constant.

Fail Safes: Snap disc thermostats, air flow sensors and thermocouples. Over tem-
perature limit of 235

°

C.

Air Dryer: Dries air to -70

°

C dew point.

Air Flow: 200 to 960 SCFH (1.6 to 7.6 liters/sec)

*

Cold specifications are for room ambient temp. of 75 F, or lower, with 60 Hz power.

Transition time is from +125 C to -55 C or -55 C to +125 C; air temperature measured at air

output nozzle. Measured in temperature cycle mode after system stabilization with a 10 second
soak time, or less, at each temperature. Room ambient temperature to be below 75

°

F

.

X

Y

R

(in/cm)

(in/cm)

(in/cm)

ARM

MAX

MIN

MAX

MIN

MAX

MIN

TF-IC5

57/145

24/61

30/76

18/46

54/137

38/96

TF-IC7

57/145

24/61

30/76

18/46

54/137

38/96

ENVIRONMENT

Ambient temperature range:

Operating:

15

°

C to 28

°

C

Non-Operating:

-40

°

C to +85

°

C

Relative Humidity: 20 to 65%

AIR INPUT REQUIREMENTS

Pressure:80-110 PSIG (5.7 to 7.8 Kg/cm

2

)

Flow Rate: 10-16 SCFM (4.7-7.6 liters/sec)

Dew Point: Less than 10C at 80

PSIG (5.7Kg/cm2)

Quality: Clean, dry air, free of oil, moisture and particles.

Optional transformers are available for input voltage ranges of 195-
210 VAC or 220-245 VAC.

FRONT VIEW

TOP VIEW

A Test Enterprises Company