Technical specifications – HT instruments T2100 User Manual

Page 26

Advertising
background image

T2000-T2100

EN – 25

6. TECHNICAL SPECIFICATIONS

6.1. REFERENCE

CONDITIONS

Parameter Reference

condition

Environmental temperature

20°C

 3°C

Relative humidity

50%RH

 10%

Battery voltage

6V

 0.5V

External magnetic field

<40A/m

External electric field

<1V/m

Clamp positioning

Horizontal

Position of the conductor in the clamp

Centred

Closeness to metal masses

> 10cm

Loop resistances

None

Measured sinusoidal current frequency

50Hz

Distortion percentage

<0.5%

Disturbance current in resistance measurement

None


6.2. TECHNICAL

CHARACTERISTICS

Accuracy is indicated as [% of reading + value] referred to the reference conditions.

Resistance

Range [

] Resolution

[

]

Accuracy

0.010

 0.099

0.001

(1%rdg + 0.01)

0.10

 0.99

0.01

1.0

 49.9

0.1

(1.5%rdg + 0.1)

50.0

 99.5

0.5

(2%rdg + 0.5)

100

 199

1

(3%rdg + 1)

200

 395

5

(6%rdg + 5)

400

 590

10

(10%rdg + 10)

600

 1000

20

(20%rdg + 20)

If measured resistance is > 1000

, the display shows “OL”

Resistance measuring frequency: >1kHz

Measuring range of resistance alarm threshold setting: 1

 999

TRMS AC Current (T2000)

Range Resolution

Accuracy

0.00mA

 80.00mA

0.05mA

(2.5%rdg + 1mA)

80.0mA

 650.0mA

0.5mA

(2.5%rdg + 2mA)

650.0mA

 999.5mA

1.5mA

(2.5%rdg + 10mA)

1.000A

 3.995A

0.005A

(2.5%rdg + 0.01A)

4.00A

 20.00A

0.01A

(2.5%rdg + 0.02A)

Current measuring frequency: 50

60Hz

Measuring range of current alarm threshold setting: 1mA

999mA

For non-sinusoidal waveforms (CF <> 1.414 and CF < 1,9) add 1.5% to percentage error