T2000-T2100

EN – 3

1.2. DURING

USE

Please carefully read the following recommendations and instructions:

CAUTION

Failure to comply with the Caution notes and/or Instructions may damage the
instrument and/or its components or be a source of danger for the operator.

 Operate the clamp lever twice before switching on the instrument, to make sure that the

clamp jaws are completely closed.

 When switching on the instrument, DO NOT operate the clamp lever and do not clamp

any cable.

 Avoid measuring resistance if external voltages are present. Even if the instrument is

protected, excessive voltage could cause malfunctions.

 During current measurement (T2000), any other current near the clamp may affect

measurement precision.

 When measuring current (T2000), always put the conductor as near as possible to the

middle of the clamp jaw, to obtain the most accurate reading.

 While measuring, if the value of the quantity being measured remains unchanged, check

if the HOLD function is enabled.

CAUTION

If the symbol “

” is displayed during use, interrupt testing, disconnect the

instrument from the system, switch off the instrument and replace its batteries
(see § 5.2).

1.3. AFTER

USE

 When measurements are completed, switch off the instrument by pressing the

multifunction key

“1” (see § 4.4).

 If the instrument is not to be used for a long time, remove the batteries.

1.4.

DEFINITION OF MEASUREMENT (OVERVOLTAGE) CATEGORY

Standard “IEC/EN61010-1: Safety requirements for electrical equipment for measurement,
control and laboratory use, Part 1: General requirements” defines what measurement
category, commonly called overvoltage category, is. § 6.7.4: Measured circuits, reads:

Circuits are divided into the following measurement categories:

Measurement category IV is for measurements performed at the source of the low-

voltage installation.
Examples are electricity meters and measurements on primary overcurrent protection
devices and ripple control units

Measurement category III is for measurements performed on installations inside

buildings.

Examples are measurements on distribution boards, circuit breakers, wiring, including

cables, bus-bars, junction boxes, switches, socket-outlets in the fixed installation, and
equipment for industrial use and some other equipment, for example, stationary motors
with permanent connection to fixed installation

Measurement category II is for measurements performed on circuits directly connected

to the low-voltage installation.
Examples are measurements on household appliances, portable tools and similar
equipment.

Measurement category I is for measurements performed on circuits not directly

connected to MAINS.

Examples are measurements on circuits not derived from MAINS, and specially protected
(internal) MAINS-derived circuits. In the latter case, transient stresses are variable; for
that reason, the standard requires that the transient withstand capability of the equipment
is made known to the user.