Chapter 8: Troubleshooting and Servicer Operating Log
Product Service 1-800-522-7658
44
44
PROBLEM
CAUSE
CORRECTIVE ACTION
Airflow monitor
malfunction
(Continued)
Wrong alarm set
point.
Airflow monitor was not properly
adjusted. Repeat calibration steps
outlined in this manual in Chapter 6.
Constant audible
alarm.
Check airflow and calibration of
airflow monitor. See Chapter 6.
Continuous alarm.
Check the downflow velocity of the
enclosure as the airflow of the
system may have changed. If
incorrect, adjust the speed control to
increase downflow velocity. The
HEPA filter may have become
loaded. If downflow velocity is
correct, calibrate the airflow monitor
outlined in this manual in Chapter 6.
Monitor alarms; air
way to airflow
monitor sensor is
blocked by insects,
dust or debris.
Lightly clean the airway with clean
air. Be careful not to touch sensitive
electrical components.
Audible disable
will not stay
operational.
An alarm condition must be
continuously present before the
audible alarm can be silenced. If
flow conditions fluctuate near the
alarm set point, the airflow monitor
will automatically reset it. Action
should be taken to bring the
enclosure airflow into proper
operating parameters or adjust the
alarm set point lower.
- Purifie Horizontal Clean Benches 3970423 Purifie Horizontal Clean Benches 3970421 Purifie Horizontal Clean Benches 3970420 Purifie Horizontal Clean Benches 3970404 Purifie Horizontal Clean Benches 3970403 Purifie Horizontal Clean Benches 3970401 Purifie Horizontal Clean Benches 3970400 Purifie Horizontal Clean Benches 3970324 Purifie Horizontal Clean Benches 3970323 Purifie Horizontal Clean Benches 3970321 Purifie Horizontal Clean Benches 3970320 Purifie Horizontal Clean Benches 3970304 Purifie Horizontal Clean Benches 3970303 Purifie Horizontal Clean Benches 3970301 Purifie Horizontal Clean Benches 3970300 Purifie Horizontal Clean Benches 3970224 Purifie Horizontal Clean Benches 3970223 Purifie Horizontal Clean Benches 3970221 Purifie Horizontal Clean Benches 3970220 Purifie Horizontal Clean Benches 3970204 Purifie Horizontal Clean Benches 3970203 Purifie Horizontal Clean Benches 3970201 Purifie Horizontal Clean Benches 3970200